DC current

I would like to know if I can use an AC amp meter to check DC current up to 300 amps. my meter is rated for 300 amps AC, I have read that it possible to do this. However I would more info about the subject before trying this.
 
Sponsored Links
  #2  
Old 09-02-18, 09:14 AM
PJmax's Avatar
Group Moderator
Join Date: Oct 2012
Location: Northern NJ - USA
Posts: 59,023
Received 1,104 Votes on 1,024 Posts
Welcome to the forums.

As far as I know.....an AC ammeter will not read DC amperage properly.
The clamp-on probes work by measuring the magnetic strength across the wire. AC has a an alternating magnetic field while DC does not. A DC clamp meter usually uses a Hall effect transistor.

You can purchase a DC shunt that will indicate the DC amperage as a DC voltage that any DC voltmeter can read. This shunt is installed in series in one of the power legs.
https://www.google.com/search?client...w=1067&bih=511
 
  #3  
Old 09-02-18, 05:10 PM
L
Member
Join Date: Nov 2015
Posts: 1,463
Received 68 Votes on 62 Posts
AC amp meter will not read DC current at all. Theoretically it should read 0A with AC current meter, but in reality it will read very low current which is a result from the noise on DC circuit.

You will need a shunt as PJmax suggested. It is basically a high current resistor with low resistant. You will be measuring the voltage across a shunt and since you know the resistance of the shunt, you can calculate the current using ohm's law.
 
  #4  
Old 09-03-18, 04:53 AM
GregH's Avatar
Super Moderator
Join Date: Jan 2001
Location: Manitoba
Posts: 10,139
Received 37 Votes on 35 Posts
A DC clamp on ammeter is available but they are more costly than an AC only type.


There are many places to purchase these.
Shop around it interested.
 
  #5  
Old 09-03-18, 05:40 AM
L
Member
Join Date: Nov 2015
Posts: 1,463
Received 68 Votes on 62 Posts
DC Clamp meter is not recommended for anything requiring accurate reading or permanent installation.
Because DC Clamp meters use hall effect, it is very sensitive to nearby magnetic field and even Earth's magnetic field affects it. All DC clamp meters require zero before measuring for this reason.
 
  #6  
Old 09-03-18, 06:28 AM
T
Member
Join Date: Oct 2005
Location: USA
Posts: 1,625
Received 41 Votes on 40 Posts
Originally Posted by lambition View Post
DC Clamp meter is not recommended for anything requiring accurate reading or permanent installation.
Because DC Clamp meters use hall effect, it is very sensitive to nearby magnetic field and even Earth's magnetic field affects it. All DC clamp meters require zero before measuring for this reason.
Funny you mention this. I recently tested a Sentran (non-split core) 20Amp hall effect for 60Hz mag field resistance. It was solid to 100A/m field strength on all axis.
Cheap clamp-on/split cores may be a different story.
 
Reply
Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off


Thread Tools
Search this Thread
 
Ask a Question
Question Title:
Description:
Your question will be posted in: